Abstract : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
author: Renzo Guarnieri, Dario Di Nardo, Gianni Di Giorgio, Gabriele Miccoli, Luca Testarelli | publisher: drg. Andreas Tjandra, Sp. Perio, FISID
To evaluate and compare radiographic crestal bone loss (CBL) and soft tissue parameters around submerged/two-stage and nonsubmerged/one-stage single implants with the same endosseous portion (body design and surface, thread design and distance) and identical intramucosal laser-microgrooved surface, after 3 years of loading.
Twenty submerged/two-stage implants and 20 nonsubmerged/one-stage implants were placed randomly with a split-mouth design in the posterior areas of 20 partially edentulous patients.
Radiographic and clinical examinations were carried out at the implant placement, at the delivery of prosthetic restorations, and at each year of the follow-up period. Plaque index (PI), probing depth (PD), bleeding on probing (BOP), and gingival recession (REC) were recorded. Radiographic crestal bone levels were assessed at the mesial and distal aspect of the implant sites. In addition, the influence of the vertical keratinized tissue thickness (KTT) on CBL was investigated.
At the delivery of prosthetic restorations, a statistically significant difference (P = 0.013) was found in radiographic mean CBL between submerged and nonsubmerged implants (0.15 ± 0.05 mm vs. 0.11 ± 0.04 mm). At the end of the follow-up period, no statistical difference (P = 0.741) was found in the mean CBL between submerged and nonsubmerged implants (0.27 ± 04 mm vs. 0.26 ± 0.5 mm). The changes in the soft tissues including PI, PD, BOP, and REC had no significant differences in either group. Moreover, KTT did not show a statistical correlation with CBL.
After 3 years of loading, no statistical difference was noted in CBL and soft tissue conditions between single submerged two-stage and nonsubmerged one-stage laser-microgrooved implants.
http://clinicaltrials.gov/ct2/show/NCT03674762
Serial posts:
-
Abstract : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
Introduction : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
Materials and methods : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [1]
-
Materials and methods : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [2]
-
Materials and methods : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [3]
-
Materials and methods : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [4]
-
Materials and methods : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [5]
-
Results : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
Discussion : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [1]
-
Discussion : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [2]
-
Discussion : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [3]
-
Conclusions : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
Availability of data and materials : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
References : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [1]
-
References : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [2]
-
References : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [3]
-
References : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [4]
-
Acknowledgments : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
Funding : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
Author information : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
Ethics declarations : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
Additional information : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
Rights and permissions : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
About this article : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
-
Table 1 Demographic data of patients, implants position, and type of implant : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single
-
Table 2 Distribution of each implant in each group : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in
-
Table 3 Differences in number of sites with plaque and bleeding on probing (BOP) between the two groups during the follow-up period (Wilcoxon signed-rank tests, P > 0.05) : Clinical and radiographics
-
Table 4 Patients’ full-mouth periodontal probing depth (FMPPD), full-mouth plaque score (FMPS), and full-mouth bleeding score (FMBS) recorded during the follow-up period : Clinical and radiographics
-
Fig. 1. Example of the location of a non-submerged implant, bone, and adjacent tooth : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
-
Fig. 2. Example of the location of a submerged implant, bone, and adjacent tooth : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
-
Fig. 3. Implants used in the present study and laser-microtextured intramucosal surface (original magnification × 800) : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
-
Fig. 4. Schematic view of radiographic measurement references : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
-
Fig. 5. Mean values of probing depth (PD) between the two groups during the follow-up period. ANOVA test P > 0.05 : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
-
Fig. 6. Mean values of gingival recession (REC) between the two groups at the end of follow-up period (3-year). ANOVA test : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
-
Fig. 7. Changes of CBL (mm) between the two groups in sites with KKT > 2 and ≤ 2 mm. ANOVA test : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
-
Fig. 8. Mean values of crestal bone loss (CBL) between the two groups during the follow-up period. ANOVA test : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
To evaluate and compare radiographic crestal bone loss (CBL) and soft tissue parameters around submerged/two-stage and nonsubmerged/one-stage single implants with the same endosseous portion (body design and surface, thread design and distance) and identical intramucosal laser-microgrooved surface, after 3 years of loading.
Twenty submerged/two-stage implants and 20 nonsubmerged/one-stage implants were placed randomly with a split-mouth design in the posterior areas of 20 partially edentulous patients.
Radiographic and clinical examinations were carried out at the implant placement, at the delivery of prosthetic restorations, and at each year of the follow-up period. Plaque index (PI), probing depth (PD), bleeding on probing (BOP), and gingival recession (REC) were recorded. Radiographic crestal bone levels were assessed at the mesial and distal aspect of the implant sites. In addition, the influence of the vertical keratinized tissue thickness (KTT) on CBL was investigated.
At the delivery of prosthetic restorations, a statistically significant difference (P = 0.013) was found in radiographic mean CBL between submerged and nonsubmerged implants (0.15 ± 0.05 mm vs. 0.11 ± 0.04 mm). At the end of the follow-up period, no statistical difference (P = 0.741) was found in the mean CBL between submerged and nonsubmerged implants (0.27 ± 04 mm vs. 0.26 ± 0.5 mm). The changes in the soft tissues including PI, PD, BOP, and REC had no significant differences in either group. Moreover, KTT did not show a statistical correlation with CBL.
After 3 years of loading, no statistical difference was noted in CBL and soft tissue conditions between single submerged two-stage and nonsubmerged one-stage laser-microgrooved implants.
http://clinicaltrials.gov/ct2/show/NCT03674762
- Abstract : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- Introduction : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- Materials and methods : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [1]
- Materials and methods : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [2]
- Materials and methods : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [3]
- Materials and methods : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [4]
- Materials and methods : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [5]
- Results : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- Discussion : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [1]
- Discussion : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [2]
- Discussion : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [3]
- Conclusions : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- Availability of data and materials : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- References : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [1]
- References : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [2]
- References : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [3]
- References : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas [4]
- Acknowledgments : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- Funding : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- Author information : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- Ethics declarations : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- Additional information : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- Rights and permissions : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- About this article : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in posterior areas
- Table 1 Demographic data of patients, implants position, and type of implant : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single
- Table 2 Distribution of each implant in each group : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implants in
- Table 3 Differences in number of sites with plaque and bleeding on probing (BOP) between the two groups during the follow-up period (Wilcoxon signed-rank tests, P > 0.05) : Clinical and radiographics
- Table 4 Patients’ full-mouth periodontal probing depth (FMPPD), full-mouth plaque score (FMPS), and full-mouth bleeding score (FMBS) recorded during the follow-up period : Clinical and radiographics
- Fig. 1. Example of the location of a non-submerged implant, bone, and adjacent tooth : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
- Fig. 2. Example of the location of a submerged implant, bone, and adjacent tooth : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
- Fig. 3. Implants used in the present study and laser-microtextured intramucosal surface (original magnification × 800) : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
- Fig. 4. Schematic view of radiographic measurement references : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
- Fig. 5. Mean values of probing depth (PD) between the two groups during the follow-up period. ANOVA test P > 0.05 : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
- Fig. 6. Mean values of gingival recession (REC) between the two groups at the end of follow-up period (3-year). ANOVA test : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
- Fig. 7. Changes of CBL (mm) between the two groups in sites with KKT > 2 and ≤ 2 mm. ANOVA test : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant
- Fig. 8. Mean values of crestal bone loss (CBL) between the two groups during the follow-up period. ANOVA test : Clinical and radiographics results at 3 years of RCT with split-mouth design of submerged vs. nonsubmerged single laser-microgrooved implant