Introduction : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
author: Waldemar Reich, Ramona Schweyen, Christian Heinzelmann, Jeremias Hey, Bilal Al-Nawas, Alexander Walter Eckert | publisher: drg. Andreas Tjandra, Sp. Perio, FISID
It is known that achievement of primary stability is one precondition for osseointegration and treatment success. There are few reports of immediate [14] and early (6 weeks) functional loading of short implants [28]. This is related to good bone quality, implant design or implant site preparation (e.g. under-drilling). However, under-drilling of the crestal aspect may lead to decreased bone-to-implant contact [29]. It is desirable to reduce the periimplant stress on the crestal bone while providing sufficient primary stability for all bone densities.
Therefore, optimisation of the macro- and microdesign of short dental implants to improve the success rate and long-term stability is preferable. In fact, elderly patients with general comorbidity should benefit from the overall short treatment time [28, 30]. As previously shown, the oral health-related quality of life is compromised during the healing period after implant insertion [31], especially when augmentation procedures are required [11]. For several reasons, the overall treatment time should be reduced in patients with atrophic alveolar ridges.
The purpose of the present study was to clinically analyse the feasibility and safety of a new short dental implant system with an expandable compressive design in the apical region. We hypothesised that the innovative expandable macrodesign of this implant provides a reliable implant success rate and ensures high implant stability in vivo.
Serial posts:
-
Abstract : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
Introduction : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
-
Introduction : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
-
Material and methods : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
-
Material and methods : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
-
Material and methods : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [3]
-
Results : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
-
Results : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
-
Discussion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
-
Discussion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
-
Discussion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [3]
-
Discussion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [4]
-
Discussion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [5]
-
Conclusion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
Abbreviations : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
-
References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
-
References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [3]
-
References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [4]
-
References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [5]
-
References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [6]
-
References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [7]
-
Acknowledgements : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
Author information : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
-
Author information : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
-
Additional files : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
Rights and permissions : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
About this article : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
Table 1 Patient recruitment : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
Table 2 Surgical treatment protocol : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
Table 3 Prosthetic treatment protocol : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
Table 4 Clinical characteristics of the study cohort : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
-
Fig. 1. a Closed short expandable dental implant (4.1 × 7 mm). The implant-abutment connection is characterised by an internal hexagon for rotation stability, combining the advantages of conical and parallel surfaces to reduce microgaps and micromovement [68]. The microthread concept and platform switching concept are implemented in the implant shoulder to reduce periimplant bone strain [53]. b Manual fixation of the expansion tool. Take note of the distance between both yellow rings. c Completion of the expansion process using the ratchet. Take note of the contact between both yellow rings. d Opened short expandable dental implant (4.1 × 7 mm). The expanded implant provides an increased bone-to-implant interface (pyramid shape) in the apical portion [54]. e Cross-section view of the implant apex. The apical expansion process is characterised by the unfolding of four wings, which are connected by four foils. D1: diameter of the closed implant. D2: diameter of the opened implant. fTop v
-
Fig. 2. Cumulative implant survival over the follow-up period. The Kaplan-Meyer diagram visualises the analysis of implant survival in the maxilla and in the mandible (log rank test, p = 0.173) over the follow-up period up to 37 months (Table 4) : Novel expandable short dental implant
-
Fig. 3. a Primary implant stability. The histogram visualises the distribution of the implant stability quotients (ISQ) for both jaws measured by resonance frequency analysis (Osstell AB, Göteborg, Sweden). b Secondary implant stability. The histogram shows the distribution of the implant stability quotients (ISQ) of osseointegrated implants. According to the measurement implant stability was classified as low with ISQ values < 60, medium with ISQ values 60–70, and high with values ISQ > 70 [34] : Novel expandable short dental implant
-
Fig. 4. a–h Prosthetic restauration—follow-up examination. Intraoral and perioral views of a rehabilitated female patient. (She asked explicitly only for implantological treatment in the mandible.) : Novel expandable short dental implant
-
Fig. 5. a Postoperative orthopantomogram. b Follow-up orthopantomogram. c Follow-up standard periapical radiogram (implants i42 and i44). d Follow-up standard periapical radiogram (implants i32 and i34) : Novel expandable short dental implant
It is known that achievement of primary stability is one precondition for osseointegration and treatment success. There are few reports of immediate [14] and early (6 weeks) functional loading of short implants [28]. This is related to good bone quality, implant design or implant site preparation (e.g. under-drilling). However, under-drilling of the crestal aspect may lead to decreased bone-to-implant contact [29]. It is desirable to reduce the periimplant stress on the crestal bone while providing sufficient primary stability for all bone densities.
Therefore, optimisation of the macro- and microdesign of short dental implants to improve the success rate and long-term stability is preferable. In fact, elderly patients with general comorbidity should benefit from the overall short treatment time [28, 30]. As previously shown, the oral health-related quality of life is compromised during the healing period after implant insertion [31], especially when augmentation procedures are required [11]. For several reasons, the overall treatment time should be reduced in patients with atrophic alveolar ridges.
The purpose of the present study was to clinically analyse the feasibility and safety of a new short dental implant system with an expandable compressive design in the apical region. We hypothesised that the innovative expandable macrodesign of this implant provides a reliable implant success rate and ensures high implant stability in vivo.
- Abstract : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- Introduction : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
- Introduction : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
- Material and methods : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
- Material and methods : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
- Material and methods : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [3]
- Results : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
- Results : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
- Discussion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
- Discussion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
- Discussion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [3]
- Discussion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [4]
- Discussion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [5]
- Conclusion : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- Abbreviations : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
- References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
- References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [3]
- References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [4]
- References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [5]
- References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [6]
- References : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [7]
- Acknowledgements : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- Author information : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [1]
- Author information : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results [2]
- Additional files : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- Rights and permissions : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- About this article : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- Table 1 Patient recruitment : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- Table 2 Surgical treatment protocol : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- Table 3 Prosthetic treatment protocol : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- Table 4 Clinical characteristics of the study cohort : Novel expandable short dental implants in situations with reduced vertical bone height—technical note and first results
- Fig. 1. a Closed short expandable dental implant (4.1 × 7 mm). The implant-abutment connection is characterised by an internal hexagon for rotation stability, combining the advantages of conical and parallel surfaces to reduce microgaps and micromovement [68]. The microthread concept and platform switching concept are implemented in the implant shoulder to reduce periimplant bone strain [53]. b Manual fixation of the expansion tool. Take note of the distance between both yellow rings. c Completion of the expansion process using the ratchet. Take note of the contact between both yellow rings. d Opened short expandable dental implant (4.1 × 7 mm). The expanded implant provides an increased bone-to-implant interface (pyramid shape) in the apical portion [54]. e Cross-section view of the implant apex. The apical expansion process is characterised by the unfolding of four wings, which are connected by four foils. D1: diameter of the closed implant. D2: diameter of the opened implant. fTop v
- Fig. 2. Cumulative implant survival over the follow-up period. The Kaplan-Meyer diagram visualises the analysis of implant survival in the maxilla and in the mandible (log rank test, p = 0.173) over the follow-up period up to 37 months (Table 4) : Novel expandable short dental implant
- Fig. 3. a Primary implant stability. The histogram visualises the distribution of the implant stability quotients (ISQ) for both jaws measured by resonance frequency analysis (Osstell AB, Göteborg, Sweden). b Secondary implant stability. The histogram shows the distribution of the implant stability quotients (ISQ) of osseointegrated implants. According to the measurement implant stability was classified as low with ISQ values < 60, medium with ISQ values 60–70, and high with values ISQ > 70 [34] : Novel expandable short dental implant
- Fig. 4. a–h Prosthetic restauration—follow-up examination. Intraoral and perioral views of a rehabilitated female patient. (She asked explicitly only for implantological treatment in the mandible.) : Novel expandable short dental implant
- Fig. 5. a Postoperative orthopantomogram. b Follow-up orthopantomogram. c Follow-up standard periapical radiogram (implants i42 and i44). d Follow-up standard periapical radiogram (implants i32 and i34) : Novel expandable short dental implant