Results : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
author: Masazumi Yoshitani, Yoshiyuki Takayama, Atsuro Yokoyama | publisher: drg. Andreas Tjandra, Sp. Perio, FISID
The distributions of occlusal force are shown in Fig. 8. In model MT67 (Fig. 8e), the occlusal force at the first premolar on the defect side was 10.0–86.5 N, which was 1.2–15.0-fold larger than that on the natural dentition side. The occlusal force at the second premolar on the defect side was 24.6–190.1 N, which was 2.6–8.3-fold larger than on the natural dentition side. The occlusal force at the TMJ on the defect side was 26.8–214.6 N, which was 1.2–1.5-fold larger than that on the natural dentition side. The occlusal force was concentrated at the second premolar on the defect side and approximately equivalent to that at the second molar on the natural dentition side.
In model MT7 (Fig. 8f), the occlusal force at the first premolar on the defect side was almost equivalent to that on the natural dentition side. The occlusal forces at the second premolar under loads 100, 200, 400, and 800 N were 8.6, 21.0, 43.2, and 84.5 N, respectively. These forces were 1.1–3.8-fold larger on the defect side than that on the natural dentition side. Additionally, the occlusal forces at the first molar under loads 100, 200, 400, and 800 N were 18.5, 47.2, 110.7, and 239.0 N, respectively. These forces were 1.5–2.1-fold larger on the defect side than that on the natural dentition side. The occlusal force at the TMJ on the defect side was 24.9–169.1 N, which was 1.1–1.3-fold larger than that on the natural dentition side. The occlusal force was concentrated at the first molar on the defect side and approximately equivalent to that at the second molar on the natural dentition side.
In model Im67 (Fig. 8a), the occlusal force at the premolars and the TMJ on the defect side was smaller than that in model MT67 (shown in Fig. 8e). The occlusal force at the second premolar and TMJ in model Im67 was 9.7–18.3 and 23.7–134.3 N, respectively. Compared with model MT67, the occlusal force at the second premolar and left TMJ in model Im67 was reduced by 0.005–0.5-fold and 0.6–0.9-fold, respectively. Under loads 100 and 200 N, the occlusal force at the implants was slightly smaller than that on the natural dentition side. Under load 800 N, the occlusal force at the first and second molars was 122.7 and 224.9 N, respectively. The occlusal force at the implants was slightly larger than that at natural teeth on the contralateral side. However, occlusal force distribution in model Im67 was considered to be almost symmetrical.
Serial posts:
-
Abstract : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
-
Background : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
-
Background : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [2]
-
Methods : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
-
Methods : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [2]
-
Methods : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [3]
-
Results : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
-
Results : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [2]
-
Results : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [3]
-
Discussion : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
-
Discussion : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [2]
-
Discussion : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [3]
-
Conclusions : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
-
References : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
-
References : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [2]
-
References : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [3]
-
References : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [4]
-
Author information : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
-
Ethics declarations : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
-
Rights and permissions : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
-
About this article : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
-
Table 1 Material properties : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
-
Fig. 1. Three-dimensional finite element model. The tooth roots and implant bodies are displayed with permeability. a Im67, b Im6, c Im4567, d Im456, e MT67, and f MT7 : Significance of mandibular molar replacement with a dental implant
-
Fig. 2. Load displacement curves of springs : Significance of mandibular molar replacement with a dental implant
-
Fig. 3. Load displacement curves of natural teeth in FE model : Significance of mandibular molar replacement with a dental implant
-
Fig. 4. Three-dimensional finite element model with natural teeth and no defect : Significance of mandibular molar replacement with a dental implant
-
Fig. 5. Distribution of occlusal force in the natural teeth model displayed in Fig.4 : Significance of mandibular molar replacement with a dental implant
-
Fig. 6. Springs for opposing teeth and TMJs and load directions. Arrows indicate loads, arrowheads indicate restricted nods, and spiral lines indicate springs : Significance of mandibular molar replacement with a dental implant
-
Fig. 7. Initializing models altering the load displacement curves of springs : Significance of mandibular molar replacement with a dental implant
-
Fig. 8. Distribution of occlusal force in models. a Im67, b Im6, c Im4567, d Im456, e MT67, and f MT7. R right TMJ, L left TMJ, 4 first premolar, 5 second premolar, 6 first molar, and 7 second molar. Numbers within circles indicate implant superstructure : Significance of mandibular molar replacement with a dental implant
The distributions of occlusal force are shown in Fig. 8. In model MT67 (Fig. 8e), the occlusal force at the first premolar on the defect side was 10.0–86.5 N, which was 1.2–15.0-fold larger than that on the natural dentition side. The occlusal force at the second premolar on the defect side was 24.6–190.1 N, which was 2.6–8.3-fold larger than on the natural dentition side. The occlusal force at the TMJ on the defect side was 26.8–214.6 N, which was 1.2–1.5-fold larger than that on the natural dentition side. The occlusal force was concentrated at the second premolar on the defect side and approximately equivalent to that at the second molar on the natural dentition side.
In model MT7 (Fig. 8f), the occlusal force at the first premolar on the defect side was almost equivalent to that on the natural dentition side. The occlusal forces at the second premolar under loads 100, 200, 400, and 800 N were 8.6, 21.0, 43.2, and 84.5 N, respectively. These forces were 1.1–3.8-fold larger on the defect side than that on the natural dentition side. Additionally, the occlusal forces at the first molar under loads 100, 200, 400, and 800 N were 18.5, 47.2, 110.7, and 239.0 N, respectively. These forces were 1.5–2.1-fold larger on the defect side than that on the natural dentition side. The occlusal force at the TMJ on the defect side was 24.9–169.1 N, which was 1.1–1.3-fold larger than that on the natural dentition side. The occlusal force was concentrated at the first molar on the defect side and approximately equivalent to that at the second molar on the natural dentition side.
In model Im67 (Fig. 8a), the occlusal force at the premolars and the TMJ on the defect side was smaller than that in model MT67 (shown in Fig. 8e). The occlusal force at the second premolar and TMJ in model Im67 was 9.7–18.3 and 23.7–134.3 N, respectively. Compared with model MT67, the occlusal force at the second premolar and left TMJ in model Im67 was reduced by 0.005–0.5-fold and 0.6–0.9-fold, respectively. Under loads 100 and 200 N, the occlusal force at the implants was slightly smaller than that on the natural dentition side. Under load 800 N, the occlusal force at the first and second molars was 122.7 and 224.9 N, respectively. The occlusal force at the implants was slightly larger than that at natural teeth on the contralateral side. However, occlusal force distribution in model Im67 was considered to be almost symmetrical.
- Abstract : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
- Background : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
- Background : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [2]
- Methods : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
- Methods : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [2]
- Methods : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [3]
- Results : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
- Results : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [2]
- Results : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [3]
- Discussion : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
- Discussion : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [2]
- Discussion : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [3]
- Conclusions : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
- References : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [1]
- References : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [2]
- References : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [3]
- References : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis [4]
- Author information : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
- Ethics declarations : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
- Rights and permissions : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
- About this article : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
- Table 1 Material properties : Significance of mandibular molar replacement with a dental implant: a theoretical study with nonlinear finite element analysis
- Fig. 1. Three-dimensional finite element model. The tooth roots and implant bodies are displayed with permeability. a Im67, b Im6, c Im4567, d Im456, e MT67, and f MT7 : Significance of mandibular molar replacement with a dental implant
- Fig. 2. Load displacement curves of springs : Significance of mandibular molar replacement with a dental implant
- Fig. 3. Load displacement curves of natural teeth in FE model : Significance of mandibular molar replacement with a dental implant
- Fig. 4. Three-dimensional finite element model with natural teeth and no defect : Significance of mandibular molar replacement with a dental implant
- Fig. 5. Distribution of occlusal force in the natural teeth model displayed in Fig.4 : Significance of mandibular molar replacement with a dental implant
- Fig. 6. Springs for opposing teeth and TMJs and load directions. Arrows indicate loads, arrowheads indicate restricted nods, and spiral lines indicate springs : Significance of mandibular molar replacement with a dental implant
- Fig. 7. Initializing models altering the load displacement curves of springs : Significance of mandibular molar replacement with a dental implant
- Fig. 8. Distribution of occlusal force in models. a Im67, b Im6, c Im4567, d Im456, e MT67, and f MT7. R right TMJ, L left TMJ, 4 first premolar, 5 second premolar, 6 first molar, and 7 second molar. Numbers within circles indicate implant superstructure : Significance of mandibular molar replacement with a dental implant