Biomechanical effects of offset placement of dental implants in the edentulous posterior mandible

Discussion : Biomechanical effects of offset placement of dental implants (3)

author: Yuta Shimura,Yuji Sato,Noboru Kitagawa,Miyuki Omori | publisher: drg. Andreas Tjandra, Sp. Perio, FISID

Thus, compressed displacement exhibited the same trend in the experimental models and FEA models. The results of both models may be reliable. When the effects of offset placement are considered, there is the concept of the load-supporting area (Fig. 15) put forth by Sato. The load-supporting area is the area surrounded by the lines connecting the implant peripheries, and if a loading point falls within this area, there is little lateral force on the implant body. In the L-offset with buccal loading and the B-offset with lingual loading conditions, where the loading points are close to the load-supporting area, there is significantly less force than there is with straight placement (P < 0.05). Compared to the experimental models, the FEA models showed 10 μm less compressed displacement, but this may be the influence of the minute perturbations caused by differences in the constraint conditions in the artificial mandible bottom. The ability to completely reproduce the behaviors of experimental models with the FEA models would produce results with better approximation.

Three-dimensional displacement in the FEA models

Similar trends were observed in the direction and magnitude of displacement between placements. Buccal loading exhibited considerable motion towards the buccal rotation/tilting of the implant bodies, and lingual loading exhibited little motion towards lingual displacement. This corresponds to the fact that there was more compressed displacement during buccal loading than during lingual loading.

Strain on the peri-implant bone

In past reports where models were used, it was reported that compressive strain in the peri-implant bone occurs in sites close to the loading side and tensile strain occurs on the side opposite to the loading side. Concentration of stress leads to bone resorption. In the experimental models in the present study as well, considerable compressive strain was observed in the loading-side peri-implant bone.

 