Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
author: Shinya Homma, Yasushi Makabe, Takuya Sakai, Kenzou Morinaga, Satoru Yokoue, Hirofumi Kido, Yasutomo Yajima | publisher: drg. Andreas Tjandra, Sp. Perio, FISID
Implant treatment was performed in accordance with the procedure recommended by the manufacturer, without bone augmentation. A healing abutment was connected to the implant bodies after insertion (implant insertion in one stage method). A total of 17 dentists (treatment experience, 5–35 years; average, 11.5 years) performed all implant treatments in this study. All dentists who performed the implant treatment in this study were specialist certified by Japan Society of Oral Implantology and had experience of more than 5 years implant treatment.
IT, ISQ, and voxel value were measured in this study. IT was measured immediately after the implant insertion using a torque wrench (GC, Tokyo, Japan).
ISQ was measured throughout the experimental period. To measure ISQ, a Smartpeg Type 21 (Osstell AB, Gothenburg, Sweden) was connected at 5 N cm to the implant body, measured using an Osstell ISQ™ (Osstell AB) three times from the buccal side. Average values were used for the evaluations. ISQ was measured immediately (0 week), and 4, 6, and 12 weeks after surgery in all cases, and also at 2 and 8 weeks after surgery where possible.
In the following cases, the implant body was excluded from the evaluation.
If motion and/or rotation was observed in the implant body.
If the bone surrounding the implant body showed absorption.
If inflammation was observed in tissue surrounding the implant.
If a mandatory ISQ measurement was not performed.
In this study, we performed X-ray image diagnosis using a multi-slice CT (MSCT) or a cone beam CT (CBCT) to confirm the healing of the bone form and volume after the tooth extraction. CT imaging equipment was different for each facility (two models of CBCT and one model of MSCT). It was difficult to make the same evaluation on voxel values obtained from different equipment. The X-ray examination performed in different two models of CBCT at 18 treatment sites and 8 treatment sites. Seven treatment sites were X-ray examined in MSCT. Therefore, bone quality was investigated at 18 treatment areas (8 in the maxilla, 10 in the mandible) on the CBCT performed under standardized conditions, as shown below.
Serial posts:
-
Abstract : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Background : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
Background : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
-
Results : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
Results : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
-
Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [4]
-
Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [5]
-
Conclusions : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Abbreviations : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
-
References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [4]
-
Author information : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
Author information : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
Ethics declarations : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Rights and permissions : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
About this article : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Table 1 Treatment area and size of implant body : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Table 2 Result of IT and ISQ : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Fig. 1. Genesio® Plus implant with Aanchor surface. Scheme of the dental implant body for the Genesio® Plus implants with Aanchor surface used. a Overview picture of Genesio® Plus implants with Aanchor surface. b Image from scanning electron microscopy. Both pictures were provided by GC Corporation. To obtain osseointegration from an early stage, the dental implant body was treated with sandblasting and acid etching from the neck to apex : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 2. The measurement of the voxel values. A case of bone quality diagnosis before treatment. Width and height of the bone were measured to select the proper size of the implant body. The selected implant body was simulated on the bone images as a symbol, and then the voxel value was calculated as described in the “Methods” section : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 3. The evaluation of the average ISQ. Time-lapse migration of average ISQ. Average ISQ of all specimens increased in a time-dependent manner (results indicated by a line). A significant difference was observed by 6 weeks after surgery : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 4. The classification of the insertion torque. All specimens classified into three groups according to insertion torque. Criteria for the classification are shown in the figure and in the “Methods” section : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 5. The comparison of ISQ values by the insertion torque. Time-lapse migration of ISQ values was compared with IT groups. Each IT group displayed similar migration. A significant difference in The ISQ was found in the low IT group after 8 weeks : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 6. The relationship between ISQ and insertion torque. Percentage of specimens showing ISQ ≥ 73 compared with groups by week. In all groups, a period of rapidly increasing percentages was observed (8–12 weeks in the low IT group, 4–6 weeks in the medium and high IT groups). In the medium and high IT Group, a statistically significant difference was observed between ISQ ≥ 73 and ISQ < 72 (P < 0.05) : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 7. The average voxel value between the maxilla and mandible. There was no difference between the maxilla (430.9 ± 211.6) and the mandible (475.6 ± 211.5) in the average voxel value. Also, no difference was found in each part : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 8. The relationship between average voxel value and insertion torque (averaged over the entire treatment area). The comparison of average voxel value among IT groups. Average voxel value was 384.0 ± 154.6 in the low IT group, 387.7 ± 147.7 in the medium IT group, and 619.2 ± 200.4 in the high IT group : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 9. The comparison of two groups at average voxel values for each part. The comparison of voxel values by insertion torque. All specimens were classified into two groups by insertion torque < 40 and ≥ 40. The < 40 group represents a combination of the low and medium IT groups : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
Implant treatment was performed in accordance with the procedure recommended by the manufacturer, without bone augmentation. A healing abutment was connected to the implant bodies after insertion (implant insertion in one stage method). A total of 17 dentists (treatment experience, 5–35 years; average, 11.5 years) performed all implant treatments in this study. All dentists who performed the implant treatment in this study were specialist certified by Japan Society of Oral Implantology and had experience of more than 5 years implant treatment.
IT, ISQ, and voxel value were measured in this study. IT was measured immediately after the implant insertion using a torque wrench (GC, Tokyo, Japan).
ISQ was measured throughout the experimental period. To measure ISQ, a Smartpeg Type 21 (Osstell AB, Gothenburg, Sweden) was connected at 5 N cm to the implant body, measured using an Osstell ISQ™ (Osstell AB) three times from the buccal side. Average values were used for the evaluations. ISQ was measured immediately (0 week), and 4, 6, and 12 weeks after surgery in all cases, and also at 2 and 8 weeks after surgery where possible.
In the following cases, the implant body was excluded from the evaluation.
If motion and/or rotation was observed in the implant body.
If the bone surrounding the implant body showed absorption.
If inflammation was observed in tissue surrounding the implant.
If a mandatory ISQ measurement was not performed.
In this study, we performed X-ray image diagnosis using a multi-slice CT (MSCT) or a cone beam CT (CBCT) to confirm the healing of the bone form and volume after the tooth extraction. CT imaging equipment was different for each facility (two models of CBCT and one model of MSCT). It was difficult to make the same evaluation on voxel values obtained from different equipment. The X-ray examination performed in different two models of CBCT at 18 treatment sites and 8 treatment sites. Seven treatment sites were X-ray examined in MSCT. Therefore, bone quality was investigated at 18 treatment areas (8 in the maxilla, 10 in the mandible) on the CBCT performed under standardized conditions, as shown below.
- Abstract : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Background : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- Background : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
- Results : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- Results : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
- Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [4]
- Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [5]
- Conclusions : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Abbreviations : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
- References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [4]
- Author information : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- Author information : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- Ethics declarations : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Rights and permissions : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- About this article : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Table 1 Treatment area and size of implant body : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Table 2 Result of IT and ISQ : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Fig. 1. Genesio® Plus implant with Aanchor surface. Scheme of the dental implant body for the Genesio® Plus implants with Aanchor surface used. a Overview picture of Genesio® Plus implants with Aanchor surface. b Image from scanning electron microscopy. Both pictures were provided by GC Corporation. To obtain osseointegration from an early stage, the dental implant body was treated with sandblasting and acid etching from the neck to apex : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 2. The measurement of the voxel values. A case of bone quality diagnosis before treatment. Width and height of the bone were measured to select the proper size of the implant body. The selected implant body was simulated on the bone images as a symbol, and then the voxel value was calculated as described in the “Methods” section : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 3. The evaluation of the average ISQ. Time-lapse migration of average ISQ. Average ISQ of all specimens increased in a time-dependent manner (results indicated by a line). A significant difference was observed by 6 weeks after surgery : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 4. The classification of the insertion torque. All specimens classified into three groups according to insertion torque. Criteria for the classification are shown in the figure and in the “Methods” section : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 5. The comparison of ISQ values by the insertion torque. Time-lapse migration of ISQ values was compared with IT groups. Each IT group displayed similar migration. A significant difference in The ISQ was found in the low IT group after 8 weeks : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 6. The relationship between ISQ and insertion torque. Percentage of specimens showing ISQ ≥ 73 compared with groups by week. In all groups, a period of rapidly increasing percentages was observed (8–12 weeks in the low IT group, 4–6 weeks in the medium and high IT groups). In the medium and high IT Group, a statistically significant difference was observed between ISQ ≥ 73 and ISQ < 72 (P < 0.05) : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 7. The average voxel value between the maxilla and mandible. There was no difference between the maxilla (430.9 ± 211.6) and the mandible (475.6 ± 211.5) in the average voxel value. Also, no difference was found in each part : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 8. The relationship between average voxel value and insertion torque (averaged over the entire treatment area). The comparison of average voxel value among IT groups. Average voxel value was 384.0 ± 154.6 in the low IT group, 387.7 ± 147.7 in the medium IT group, and 619.2 ± 200.4 in the high IT group : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 9. The comparison of two groups at average voxel values for each part. The comparison of voxel values by insertion torque. All specimens were classified into two groups by insertion torque < 40 and ≥ 40. The < 40 group represents a combination of the low and medium IT groups : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant