Results : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
author: Shinya Homma, Yasushi Makabe, Takuya Sakai, Kenzou Morinaga, Satoru Yokoue, Hirofumi Kido, Yasutomo Yajima | publisher: drg. Andreas Tjandra, Sp. Perio, FISID
A total of 33 implant bodies (8 in the maxilla, 25 in the mandible) were inserted into the 27 participants (11 men, 16 women), with the average age of 54.6 ± 12.2 years (range, 32–78 years). The average IT value was 32.7 ± 9.2 N cm (32.5 ± 11.6 N cm in the maxilla, 32.8 ± 8.5 N cm in the mandible). The diameter of the implant body was 4.4 mm in 20 (60.6%) and 3.8 mm in 13 (39.4%), and the length of the implant body was 8.0 mm in 6 (18.2%), 10.0 mm in 19 (57.6%), and 12.0 mm in 8 (24.2%) (Table 1).
The measurement results of IT and ISQ are shown in Table 2. Due to the identification of mobility at 4 weeks postoperatively, No. 19 implant body (diameter, 4.4 mm; length, 12.0 mm) that had been inserted at the first molar position in the right mandible of a 32-year-old male patient was excluded from the evaluation. As a result, this study evaluated the 32 implant bodies (8 in the maxilla, 24 in the mandible) inserted into 26 participants. The survival rate of the implant bodies at the end of this study was 97% (maxilla, 100%; mandible, 96%).
The average ISQ values of all tested implant bodies increased through this study; moreover, all tested implant bodies indicated 60 or more ISQ value 6 weeks after the implant insertion. A significant difference was observed at 0 and ≥ 6 weeks (P < 0.01) (Fig. 3). No significant difference of average ISQ was found on the maxilla and mandible.
The inserted implant bodies were classified by IT value as the low IT group (< 30 N cm), the medium IT group (30–40 N cm), and the high IT group (≥ 40 N cm). Nine specimens were classified as the low IT group (3 in the maxilla, 6 in the mandible), 12 as the medium IT group (1 in the maxilla, 11 in the mandible), and 11 as the high IT group (4 in the maxilla, 7 in the mandible) (Fig. 4). There was no difference between the maxilla and the mandible in the average value of the IT (maxilla 32.5 ± 11.6 N cm, mandible 32.9 ± 8.7 N cm).
Serial posts:
-
Abstract : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Background : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
Background : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
-
Results : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
Results : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
-
Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [4]
-
Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [5]
-
Conclusions : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Abbreviations : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
-
References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [4]
-
Author information : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
-
Author information : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
-
Ethics declarations : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Rights and permissions : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
About this article : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Table 1 Treatment area and size of implant body : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Table 2 Result of IT and ISQ : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
-
Fig. 1. Genesio® Plus implant with Aanchor surface. Scheme of the dental implant body for the Genesio® Plus implants with Aanchor surface used. a Overview picture of Genesio® Plus implants with Aanchor surface. b Image from scanning electron microscopy. Both pictures were provided by GC Corporation. To obtain osseointegration from an early stage, the dental implant body was treated with sandblasting and acid etching from the neck to apex : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 2. The measurement of the voxel values. A case of bone quality diagnosis before treatment. Width and height of the bone were measured to select the proper size of the implant body. The selected implant body was simulated on the bone images as a symbol, and then the voxel value was calculated as described in the “Methods” section : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 3. The evaluation of the average ISQ. Time-lapse migration of average ISQ. Average ISQ of all specimens increased in a time-dependent manner (results indicated by a line). A significant difference was observed by 6 weeks after surgery : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 4. The classification of the insertion torque. All specimens classified into three groups according to insertion torque. Criteria for the classification are shown in the figure and in the “Methods” section : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 5. The comparison of ISQ values by the insertion torque. Time-lapse migration of ISQ values was compared with IT groups. Each IT group displayed similar migration. A significant difference in The ISQ was found in the low IT group after 8 weeks : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 6. The relationship between ISQ and insertion torque. Percentage of specimens showing ISQ ≥ 73 compared with groups by week. In all groups, a period of rapidly increasing percentages was observed (8–12 weeks in the low IT group, 4–6 weeks in the medium and high IT groups). In the medium and high IT Group, a statistically significant difference was observed between ISQ ≥ 73 and ISQ < 72 (P < 0.05) : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 7. The average voxel value between the maxilla and mandible. There was no difference between the maxilla (430.9 ± 211.6) and the mandible (475.6 ± 211.5) in the average voxel value. Also, no difference was found in each part : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 8. The relationship between average voxel value and insertion torque (averaged over the entire treatment area). The comparison of average voxel value among IT groups. Average voxel value was 384.0 ± 154.6 in the low IT group, 387.7 ± 147.7 in the medium IT group, and 619.2 ± 200.4 in the high IT group : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
-
Fig. 9. The comparison of two groups at average voxel values for each part. The comparison of voxel values by insertion torque. All specimens were classified into two groups by insertion torque < 40 and ≥ 40. The < 40 group represents a combination of the low and medium IT groups : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
A total of 33 implant bodies (8 in the maxilla, 25 in the mandible) were inserted into the 27 participants (11 men, 16 women), with the average age of 54.6 ± 12.2 years (range, 32–78 years). The average IT value was 32.7 ± 9.2 N cm (32.5 ± 11.6 N cm in the maxilla, 32.8 ± 8.5 N cm in the mandible). The diameter of the implant body was 4.4 mm in 20 (60.6%) and 3.8 mm in 13 (39.4%), and the length of the implant body was 8.0 mm in 6 (18.2%), 10.0 mm in 19 (57.6%), and 12.0 mm in 8 (24.2%) (Table 1).
The measurement results of IT and ISQ are shown in Table 2. Due to the identification of mobility at 4 weeks postoperatively, No. 19 implant body (diameter, 4.4 mm; length, 12.0 mm) that had been inserted at the first molar position in the right mandible of a 32-year-old male patient was excluded from the evaluation. As a result, this study evaluated the 32 implant bodies (8 in the maxilla, 24 in the mandible) inserted into 26 participants. The survival rate of the implant bodies at the end of this study was 97% (maxilla, 100%; mandible, 96%).
The average ISQ values of all tested implant bodies increased through this study; moreover, all tested implant bodies indicated 60 or more ISQ value 6 weeks after the implant insertion. A significant difference was observed at 0 and ≥ 6 weeks (P < 0.01) (Fig. 3). No significant difference of average ISQ was found on the maxilla and mandible.
The inserted implant bodies were classified by IT value as the low IT group (< 30 N cm), the medium IT group (30–40 N cm), and the high IT group (≥ 40 N cm). Nine specimens were classified as the low IT group (3 in the maxilla, 6 in the mandible), 12 as the medium IT group (1 in the maxilla, 11 in the mandible), and 11 as the high IT group (4 in the maxilla, 7 in the mandible) (Fig. 4). There was no difference between the maxilla and the mandible in the average value of the IT (maxilla 32.5 ± 11.6 N cm, mandible 32.9 ± 8.7 N cm).
- Abstract : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Background : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- Background : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- Methods : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
- Results : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- Results : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
- Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [4]
- Discussion : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [5]
- Conclusions : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Abbreviations : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [3]
- References : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [4]
- Author information : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [1]
- Author information : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region [2]
- Ethics declarations : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Rights and permissions : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- About this article : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Table 1 Treatment area and size of implant body : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Table 2 Result of IT and ISQ : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implants in the posterior region
- Fig. 1. Genesio® Plus implant with Aanchor surface. Scheme of the dental implant body for the Genesio® Plus implants with Aanchor surface used. a Overview picture of Genesio® Plus implants with Aanchor surface. b Image from scanning electron microscopy. Both pictures were provided by GC Corporation. To obtain osseointegration from an early stage, the dental implant body was treated with sandblasting and acid etching from the neck to apex : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 2. The measurement of the voxel values. A case of bone quality diagnosis before treatment. Width and height of the bone were measured to select the proper size of the implant body. The selected implant body was simulated on the bone images as a symbol, and then the voxel value was calculated as described in the “Methods” section : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 3. The evaluation of the average ISQ. Time-lapse migration of average ISQ. Average ISQ of all specimens increased in a time-dependent manner (results indicated by a line). A significant difference was observed by 6 weeks after surgery : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 4. The classification of the insertion torque. All specimens classified into three groups according to insertion torque. Criteria for the classification are shown in the figure and in the “Methods” section : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 5. The comparison of ISQ values by the insertion torque. Time-lapse migration of ISQ values was compared with IT groups. Each IT group displayed similar migration. A significant difference in The ISQ was found in the low IT group after 8 weeks : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 6. The relationship between ISQ and insertion torque. Percentage of specimens showing ISQ ≥ 73 compared with groups by week. In all groups, a period of rapidly increasing percentages was observed (8–12 weeks in the low IT group, 4–6 weeks in the medium and high IT groups). In the medium and high IT Group, a statistically significant difference was observed between ISQ ≥ 73 and ISQ < 72 (P < 0.05) : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 7. The average voxel value between the maxilla and mandible. There was no difference between the maxilla (430.9 ± 211.6) and the mandible (475.6 ± 211.5) in the average voxel value. Also, no difference was found in each part : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 8. The relationship between average voxel value and insertion torque (averaged over the entire treatment area). The comparison of average voxel value among IT groups. Average voxel value was 384.0 ± 154.6 in the low IT group, 387.7 ± 147.7 in the medium IT group, and 619.2 ± 200.4 in the high IT group : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant
- Fig. 9. The comparison of two groups at average voxel values for each part. The comparison of voxel values by insertion torque. All specimens were classified into two groups by insertion torque < 40 and ≥ 40. The < 40 group represents a combination of the low and medium IT groups : Prospective multicenter non-randomized controlled study on intraosseous stability and healing period for dental implant