Discussion : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [4]
Several studies have been conducted on this type of implant–abutment connection [23, 34]. Sutter et al. [35] demonstrated reverse torque values of this hybrid implant–abutment connection that were 124% of the initial tightening torque. These authors suggested that cold welding occurred in the conical hybrid implant–abutment connection.
Schmitt et al. [36] compared conical and nonconical implant–abutment connection systems in terms of their in vitro and in vivo performances. In vitro studies indicate that conical and nonconical abutments exhibited sufficient resistance to the maximal bending forces and fatigue loading. However, conical abutments were superior in terms of sealing, microgap formation, torque maintenance, and abutment stability.
According to the results of this study, %RTL increased significantly with increasing collar length and the percentage of removal torque gaining after load with 0° abutments with collar height 4 mm was less than that of 2 mm. This finding matches what was found by Siadat et al. [9] who evaluated the effect of collar length on screw loosening and concluded that an increase in the abutment collar length significantly increase the torque loss of abutment–implant screw after cyclic loading. The abutment collar length acts as vertical cantilever, so increasing abutment collar would lead to an increased vertical cantilever which acts as a force magnifier [8]. Cantilever designs increase bending forces on screws due to the lever effect [17, 18].
Serial posts:
- Abstract : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading
- Introduction : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [1]
- Introduction : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [2]
- Introduction : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [3]
- Introduction : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [4]
- Materials and methods : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [1]
- Materials and methods : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [2]
- Materials and methods : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [3]
- Results : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading
- Discussion : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [1]
- Discussion : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [2]
- Discussion : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [3]
- Discussion : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [4]
- Conclusions : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading
- References : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [1]
- References : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [2]
- References : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading [3]
- Acknowledgements : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading
- Author information : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading
- Ethics declarations : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading
- Rights and permissions : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading
- About this article : Effect of different angulations and collar lengths of conical hybrid implant abutment on screw loosening after dynamic cyclic loading
- Table 1 One-way ANOVA and post hoc Tukey test results for mean ± SD of the %initial RTL, %postload RTL, and %difference between initial and postload RTL between all groups (Of: Effect of different angulations and collar lengths of conical hybrid implant)
- Table 2 One-way ANOVA and post hoc Tukey test results for mean ± SD of the initial RTV, postload RTV, and difference between initial and postload RTV between all groups (Of: Effect of different angulations and collar lengths of conical hybrid implant)
- Table 3 Comparison between short and high collar length (A and B) (Of: Effect of different angulations and collar lengths of conical hybrid implant)
- Table 4 The raw data in all six experimental groups (Of: Effect of different angulations and collar lengths of conical hybrid implant)
- Fig. 1. Different abutment angulations and collar lengths : Effect of different angulations and collar lengths of conical hybrid implant
- Fig. 2. a Stainless steel split cylindrical mold with implant fixture screwed to abutment. b Implant fixture unscrewed from abutment after polymerization. c Implant fixture centralized vertically and perpendicular to the base with platform flushed with resin block level : Effect of different angulations and collar lengths of conical hybrid implant
- Fig. 3. 3D scanning for abutment and designing for metal tube : Effect of different angulations and collar lengths of conical hybrid implant
- Fig. 4. Application of cyclic loading with universal testing machine : Effect of different angulations and collar lengths of conical hybrid implant
- Fig. 5. Mean rate ± SD of removal torque loss (%) between groups and results of ANOVA test for loss ratio of removal torque value between groups : Effect of different angulations and collar lengths of conical hybrid implant
- Fig. 6. Comparison between short and high collar length (A and B) : Effect of different angulations and collar lengths of conical hybrid implant